Where should I write if a found a potential error in the documentation of a NI product?

I found some errata in the documentation for the PXI 2521. What is the best way to submit my findings?

www.vartortech.com
0 Kudos
Message 1 of 3
(1,908 Views)

Well, I am going to write my findings here. Hopefully it will reach the right person within NI:

 

For the latest NI PXI-2521 specifications, found here: http://www.ni.com/pdf/manuals/373797a.pdf
Page 6
Figure 2. NI PXI-2521 Connector Pinout
NOTE: This errors are also in the manual online: http://www.ni.com/documentation/en/pxi-relay-module/latest/pxi-2521/pinout/

CH2+ is missing. Instead, COM2+ repeats twice.
COM24- is missing. Instead, CH24- repeats twice.

The following error is found only in the PDF file:

CH12+ is missing. Instead, it reads C12+

www.vartortech.com
0 Kudos
Message 2 of 3
(1,879 Views)

Hello Enrique

 

Thank you very much for the feedback. To make these reports we recommend creating a service request so it can be evaluated and reported by a Technical Support Engineer.

 

For this one, I already created a Corrective Action Request (#735581) so we can work internally on fixing this issue.

 