03-15-2018 03:06 PM
So, I'm about to begin calibrating this PXIe-4136 using Calibration Executive. The card is installed in our PXIe-1062Q mainframe and is recognized in the NI MAX program. Shortly after powering on the mainframe the access light alternates between green and amber then the red voltage light illuminates and I get the following message in NI MAX:
"Status code -235115
An internal self test has failed, If this error persists, the safety of the interlock feature may be compromised. Please contact NI technical support."
This appears when I try to reset, test, or self-calibrate the device in NI MAX. The safety interlock is installed properly on the front panel and after a visual inspection of the card and interlock everything seems to be as it should be. I've calibrated multiple 4136 cards before and haven't had this issue.
Anyone have any possible guidance?
Thanks,
Joe
03-16-2018 11:38 AM
Hi ShadyJN,
It will be helpful to get a more complete understanding of your current system to help narrow down where the problem might be coming from.
If going through the safety interlock test in step 5 above fails, your PXIe-4136 SMU may be damaged. Please reach out to an Applications Engineer at ni.com/support or by calling in at (866) 275-6964 and I or one of my colleagues will be able to assist you with some further troubleshooting.
Thanks,
Jason O.
Applications Engineer
National Instruments