01-17-2019 12:53 PM
hello,
I was wondering if NI's PXIe SMU line has the option of validating remote sense continuity. Keithley SMUs refer to this as a "contact check" where the resistance is evaluated from "Force High to Sense High" and again from "Force Low to Sense Low", thereby verifying that the 4-wire connection exists at the Load and that the resistance is below a desired threshold.
01-18-2019 10:53 AM - edited 01-18-2019 10:55 AM
Hi proland1121,
You can read more about our SMUs' remote sensing capabilities here.
01-18-2019 11:43 AM - edited 01-18-2019 11:44 AM
Hi Lindsey,
Thanks for the reply but that general information page doesn't provide the level of detail required to answer the question. My main concern is the section which states "Configuring a channel for remote sense operation without connecting the sense leads to the DUT can result in measurements that do not meet the published specifications. If a channel is configured for remote sense and the remote sense leads are left open, the channel may source a voltage higher than the voltage level or voltage limit.
If I am using a probe card with 4 pins and the pin for Sense High becomes bent due to wear,operator error, or manufacturing error, the situation of "sense lead left open" occurs. Is there a way to detect this scenario? Measuring resistance between the Force/Sense pairs would identify such a scenario, but I haven't seen any such function in the NI-DC Power functions.
If no such function exists, I think i'll probably have to add in a switch matrix and a DMM (to measure resistance) for performing periodic checks.
01-21-2019 04:30 PM
To my knowledge, the NI-DC Power driver doesn't have any built-in functions that monitor those connections continuously. The switch + DMM application you mentioned seems like a good alternative.
01-26-2021 11:33 AM - edited 01-26-2021 11:35 AM
Some NI SMUs (including PXIe-4135 and PXIe-4137) have this capability. It's called sense lead error detection.
Matthew Curtis
Applications Engineer
National Instruments