PXIe-5413 maximum DC Offset

Solved!
Go to solution

Hi!
I upgraded my PXI system from the PXI-5402 waveform generator to the new dual channel  AWG PXIe-5413. Unfortunately I discovered that the DC offset for this last is only 50% of amplitude despite a more wide range of the PXI-5402:

Spec PXI-5402
DC offset PXI-5402DC offset PXI-5402

Spec PXIe-5413DC offset PXIe-5413DC offset PXIe-5413

 

I'm really surprised because it's a big limitation. For example I need a sine wave of 20 mV amplitude and a DC offset ranging from 0 to 1V. With PXI-5402 it's a simple task using standard waveform settings but with PXIe-5413 probably I have to move to arbitrary settings and create the offset digitally with "niFgen Create Waveform".
Any suggestion is appreciated!
Thanks
Lucio 

0 Kudos
Message 1 of 3
(3,117 Views)
Solution
Accepted by LucLuc

Hello LucioCina,

unfortunately your analysis is correct. With the 5413 you can't generate a 20mV amplitude sine wave with 1 V offset. You must use the arbitrary generation to achieve that goal.

 

Message 2 of 3
(3,082 Views)

Dear Lamber,

thanks for the reply despite no good news 😉
So I can do only this I suppose:
AWG offset generationAWG offset generation

For shure it'll be noiser respect the full analog offset of PXI-5402.
If you have other suggestion in mind ....i'm a big ear.

Best

0 Kudos