04-21-2022 11:29 AM
How do I clear the calibration EEPROM that stores calibration constants that have portions that are user accessible in the PXI-6713. The NI Letter of volatility says “method of clearing via API." What is the API that must be used to do this ? letter of volatility is attached, Thanks
Solved! Go to Solution.
04-21-2022 09:20 PM
You can clear the user data in calibration EEPROM by writing empty string to this property node
Any reason you want to clear the calibration constants?
04-28-2022 11:35 AM
Thank you so much