10-26-2012 07:52 AM
I calculated uncertainty measurements for a PXI-6259 card. How can I add the measurements to a report done in calibration executive?
10-29-2012 01:26 PM
Hi Df1904,
In the Calibration Executive Help file linked below
http://digital.ni.com/manuals.nsf/websearch/81A063AF6BDC757386257A100063E8E6
there is a section on Measurement Uncertainty. This section says
If you use instruments that do not have uncertainty files provided by Calibration Executive, you can create your own uncertainty files by using the template files. Template files named Calibrator Template.txt and DMM Template.txt are provided in their respective directories under Calibration Executive\Uncertainty.
10-29-2012 01:36 PM - edited 10-29-2012 01:37 PM
Hello Sunaina K.,
I do not have the Template files you are describing. Where would I find them?
10-30-2012 10:33 AM
Hi Df1904,
The template files can be found in the file path ...\Calibration Executive\Uncertainty\Templates
10-30-2012 11:01 AM
The files are not in the location
10-31-2012 03:22 PM
Hi Df1904,
Can you take a screenshot of your directory? It should look something like this:
Travis Ann
11-01-2012 08:03 AM
Here is the screen shoot. I noticed that I do not have the DMM'S and Calibrator Files.
11-02-2012 07:45 AM
Try running a repair of your Calibration Executive installation. Typically, a repair will replace files that may have not been installed when you initially installed the software.
Travis Ann
11-07-2012 07:42 AM
The files did not restore