10-08-2010 07:44 AM
When running the PXI-5610 adjustment procedure in Calibration Executive , the main power suddenly shut off . then what problem will happen to the PXI 5610?
10-12-2010 05:04 AM
What causes the main power to sunddently shut off?In general,it will not lead to any damage to PXI-5610 except great static electricity.
Please note: according to NI 5610 Calibration Procedure manual,the NI 5610 upconverter module must be cooled to 23 ±5 °C before the adjustment stage of the calibration process,and this may take up to one hour.
Thanks,
NI AE Emily
10-12-2010 07:43 AM
Thanks a lot for your reply.
After the PXI controller 's suddenly power off when performing adjustment , we restart the OS and found the PXI 5610's error infomation in the MAX : Error -200437,External calibration constants are invalid,and he card cannot be used in our Auto-Test-System anymore. I guess the calibration data in the EEPROM was missed or uncompleted,but i have no relevant softwre tools to read or write PXI-5610's EEPROM.
Could you give me some suggestions to solve this BIG problem? Thanks.