06-08-2007 01:41 PM
06-11-2007 06:41 PM
06-12-2007 12:48 PM
Here is a file of the output for the last 5610 calibration that we did. As you can see, it failed most of the frequencies and also the internal calibration frequency.
Richard
Standards used for Calibration
TypeTracking NumberCalibration Due DateNotes
Agilent ESG Family Signal Generators39804 6/7/2008
age441x 42168 3/20/2008
E9304A 34309 5/21/2008
E9304A 36165 2/19/2008
hp5313xa 28861 5/24/2008
NI 5122 38127 4/25/2009
PXI-5421 35154 11/23/2007
06-12-2007 12:50 PM
And yes, we always use the same 5421 AWG. We also installed the Cal Exec 3.2.1 upgrade and had the same issues.
Richard
06-12-2007 09:28 PM - edited 06-12-2007 09:28 PM
Hello Richard,
The calibration of the 5610 is a delicate setup and requires an excellent environment to make sure the measurements are made correctly. There are a couple of things I would suggest:
- Make sure you have the latest RFSG driver (1.2.5) for greatest accuracy corrections.
- Perform calibrations in shorter periods (six months) depending on the applications needs.
- Call or email National Instruments support for sending one of your cards here and validate the setup either with our internal lab or external. (ni.com/support).
The high percentage is just not normal on 5610 as it is not in any other NI product.
Message Edited by Yardov on 06-12-2007 09:29 PM
07-01-2008 12:08 PM
It turns out that the NI-RFSG driver needed to be updated to ver 1.3. Prior to that version, there were problems below 10MHz. The cal constants were not updated prior to verification.
See this link for a description of the problem. Even this version was released in September, we did not find out about it until December. Depending on where you go on the NI website, it might tell you that version 1.2.5 was the latest release which delayed our correction of this problem.
http://digital.ni.com/public.nsf/allkb/A3835866AC352083862572C80077FB89