The PXI-4110 current settings can only be set down to 2% of their range and if Overranging is enabled then this is extended to 1% of range.

Channel 0 of the supply has a 1A range, while Channels 1 and 2 have both 1A and 20mA ranges. This would explain why the lowest current setting on Channel 0 is higher than what you see on Channels 1 and 2. 

Steve B

0 Kudos
Message 11 of 13
(1,393 Views)

Thanks for the info. I didn't see that in the documentation I had. Do You mind telling me where you found that information.

0 Kudos
Message 12 of 13
(1,390 Views)

Never mind. I go my answer on the documentation on another post.

 