10-21-2019 11:19 AM
I am trying to calibrate the PXI-4071 DMM using the LabVIEW Calibration procedures. I'm currently going through the process of adjusting the DC Current and I am able to make it through most of the process but on the last step of calling the Cal Adjust Misc VI I get an error saying "This Calibration step has been performed out of order.". I've tried restarting the procedure a few times and I'm still getting the error near the last few steps. Does anyone know if I'm overlooking something or if there is an updated procedure that I need?
Thanks,
-Sam
05-23-2021 08:51 PM
Im having this trouble too. Did you ever find the fix for this?
05-24-2021 07:05 AM
No, I spent a few days trying to troubleshoot anything that might be throwing an error and in the end I just had to send it to NI for calibration. Wish I had a better resolution for you but that's all I have.
05-31-2021 12:30 AM
FIXED.
The problem was that i was only adjusting Current. Theres an adjustment flowchart that says you have to adjust DCV and Resistance first before adjusting current.