Dear colleagues!

Please help me, I didn't understand, with an empty and serviceable PXI-1031 the fan should work or not? In other words, not working fan with an empty PXI-1031 is talking about that PXI-1031 should be sent for repair?

The PXI-1031 is working with PXI-8106RT and I don't know what exactly caused the fault -  1031 or 8106.  

 

0 Kudos
Message 1 of 5
(2,191 Views)

Hello,

 

The fan should be working with or without modules inside it. Does it not move even with the high fan speed? If this is the case you can either send it in for repair or fix it with a fan replacement kit.

Marina B.
Technical Support Engineer
National Instruments
www.ni.com/support
0 Kudos
Message 2 of 5
(2,150 Views)

Thank you for your reply dear Marina!

Fan movement is in any condition. Does it mean that a falt is located in PXI-8106 ?

Tell me please what can replace outdated PXI-8106 ? 

0 Kudos
Message 3 of 5
(2,145 Views)

Hello,

 

Without the PXI-8106 does the fan move in your PXI-1031 when you set the speed to high? Or is it only when you have your PXI-8106 in the chassis that the fan will not move?

 

If you are thinking about purchasing a new controller I would call into our Sales Engineers. They have the skills and tools to help you spec out a new system, their number to call is (888) 280-7645.

Marina B.
Technical Support Engineer
National Instruments
www.ni.com/support
0 Kudos
Message 4 of 5
(2,112 Views)
Solution
Accepted by topic author Curr93

Hello dear Marina!

 

Without the PXI-8106 does the fan move in your PXI-1031 when you set the speed to high?

Yes

Or is it only when you have your PXI-8106 in the chassis that the fan will not move?

Fan has moving in the case with PXI-8106. Actually we localized the problem today! 

We found a faulty memory module in PXI-8106. Module size is 1024 MB. Almost 9 years of continuous operation 24/7 were not wasted. Now chassis and PXI-8106 is running. Thank you and me 🙂

0 Kudos
Message 5 of 5
(2,109 Views)