01-08-2019 04:29 PM - edited 01-08-2019 04:34 PM
I have a rack mount NI PC - RMC-8356 with a PCIe-8398 running Windows 10 with NI PXI Platform Services v18 installed
This boots fine!
I also have a NI PXIe-1085 PXI rack with a PXIe-8398 installed in slot 1
If I plug the cables from the PXIe-1085 into the PCIe-1085, the system will no longer boot (not even to the firmware).
I have double-triple confirmed that the cables are plugged in correctly and firmly.
The behavior is very consistent.
What do I do next?
Solved! Go to Solution.
01-09-2019 10:51 AM
I have heard some PCs have issues enumerating the PXI boot info. Is this an issue for the NI PC.
See http://www.ni.com/tutorial/53268/en/
01-10-2019 06:42 AM
Hi, have you tried updating the firmware?
01-15-2019 06:23 PM - edited 01-15-2019 06:46 PM
It is a NI PC and NI confirmed it is running the correct firmware.
I am now able to boot (though nothing has changes); but, I still can't talk to the PXI rack.
The good news is that NI has been able to reproduce some of the behavior and they are investigating.
Once I have a resolution, I will post it here.
03-07-2019 05:08 PM
The problem was caused by issues enumerating the PXI boot info - see http://www.ni.com/tutorial/53268/en/
It was resolved by installing NI MXI-Express BIOS Compatibility Software 18.0
http://www.ni.com/download/mxi-express-bios-compatibility-software-18.0/8020/en
Important Note: You MUST install the 18.0 or newer version it you are using WIndows 10 and/or UEFI/GPT
Previous versions can corrupt your Hard Drive and brick your PC. See https://forums.ni.com/t5/PXI/MXI-Express-BIOS-Compatibility-Software-1-5-Bricked-my-BIOS/td-p/389973...