11-07-2016 03:57 PM
Hi,
I am using the niFGEN Configure Output Mode VI to allow for two 5412 FGENs to produce waveforms with a specified phase difference. However, I am finding that whenever I run the Configure Output Mode VI, I get an error if I choose "Script" as the output mode. The error has error code -1074135039 and reads:
"niFgen Configure Output Mode.vi<ERR>Instrument error detected. Call the PREFIX_error_query function and examine the error code and error message parameters for additional information regarding the error."
Is the 5412 capable of script mode? If so, what might I be doing wrong. (If so, I will be happy to post more info).
Thanks,
Billy Maier
Solved! Go to Solution.
11-07-2016 06:12 PM
Look here: http://www.ni.com/pdf/products/us/2004_4600_305_101_DLR.pdf
Under "output modes," only " Arbitrary waveform mode and arbitrary sequence mode" are listed. So the 5412 is not capable of script mode. Compare to http://www.ni.com/datasheet/pdf/en/ds-235, for example.