I'm trying to configure a system for a new project.  One of the signals I need to measure is 3.3V, however, the signal also produces just over 6 Amp for 20 W of power. Can the PXI DAQ modules handle it?  Or even the PXI DMM module?

0 Kudos
Message 1 of 4
(2,203 Views)

Can you tell us more about what modules you are using and why you need to supply that current directly into the module?

Chase
NI Technical Support Engineer
0 Kudos
Message 2 of 4
(2,184 Views)

I have not selected a measurement module yet. One of the output voltages I need to measure, is 3.3 V.  This output also produces 6 to 7 Amps of current.  I am not interested in actually measuring the current as much as the voltage.  The only modules I've found are only rated for 2 Amps or less.

0 Kudos
Message 3 of 4
(2,172 Views)

There are no NI modules that can take that much current; you will need to transform the current down or make a voltage divider. 

Chase
NI Technical Support Engineer
0 Kudos
Message 4 of 4