Issue with Card Functionality in Rightmost Slots of PXIe-1085 Chassis

Solved!
Go to solution

Hello,

We are experiencing issues with our NI PXI setup.  Initially, we observed this on PXI-4065 DMM card.  When this DMM card is plugged into slot 12 of our PXIe-1085, our application software errors out when attempting to make a resistance measurement.  We have also noticed that in NI MAX, the DMM in slot 12 is reported as two separate devices.  One device shows up as “NI PXI-4065” with a red X and a status code of -233900 (“Internal hardware error in DMM hardware.  Please contact National Instruments support.”).  It has buttons like reset and soft panel available.  Opening the soft panel throws an error of -1074135040 “Unrecoverable failure”.  A second device shows up as “PXI-4065”, with no red X and with no buttons like reset or soft panel.

 

We think it is a PC issue because we have a duplicate NI chassis setup that communicates with a different PC (older motherboard).  When we swap the NI chasses between the PCs, the issue follows the PC, not the NI chassis.

 

Reinstalling Windows and only the NI drivers (version 17.0 May 2017 for NI-DMM, NI-PXI Platform, etc) on this problem PC does not change our DMM issues in NI Max.

 

Additionally, when the DMM is moved to a different slot (e.g., slot 2 or 9), it functions correctly in NI MAX and its soft panel opens properly.  It appears that a PXI-6133 card plugged into slot 12 also fails to function properly.

 

Could you please provide insight into what might cause issues on these slots with this PC?  Any suggestions for next steps would be greatly appreciated.

 

Hardware Configuration:

  • Chassis: NI PXIe-1085
  • Motherboard: Kontron F5110-V220
  • Operating System:  Windows 10
  • Remote Controller PCIe Card:  PCIe-8371
  • PXI Cards:
    • PXIe-8370 (MXI Express)
    • PXI-4065 (DMM)
    • PXI-2503 (MUX)
    • PXI-6133 (DAQ - Qty 😎

Below is a summary of the troubleshooting steps we have performed:

Troubleshooting Step / Observation

Result / Notes

Reboot PC and PXI chassis

Issue not resolved.

Reinstall Windows and NI drivers

No change to behavior.

Swap DAQ chassis and DMM card with a working system

The issue followed the PC, indicating the problem is not with the PXI chassis or DMM card.

Swap PCIe card with a working system

The issue remained with the problem PC.

Check "Disable Windows fast startup" setting

Already disabled in Power Options and via group policy.

Verify NI services at boot

NI Device Loader and all other NI services are running and set to automatic start.

Swap DMM (slot 12) with a DAQ card (slot 2)

The DMM card worked correctly in slot 2, but the DAQ card did not work in slot 12.

Swap PXI cable with a known-good cable

The error persisted.

Disable PC sleep/hibernate and PCI Express Link State Power Management in Windows

No change in behavior.

Verify driver versions

DMM driver version is identical between the problem system and a working system.

Run application with administrative privileges

The application still returns an error when attempting communication with DMM.

Reset Configuration Data in NI MAX

Cards in the rightmost slots of the chassis still do not function correctly.

Disable ASPM for all PCI slots

Cards in the rightmost slots still fail to function.

 

0 Kudos
Message 1 of 2
(683 Views)
Solution
Accepted by topic author jerdizzle

This was resolved using NI MXI-Express BIOS Compatibility Software