Input Resistance PXIe 6378
i have the PXIe-6378 Module and i want to know which input resistance the module has for a volatage measurement.
So i would know, how high the current is, that flows into the module.
Can you help me here?
(1,217 Views)
Accepted by topic author mihamm
Could you please elaborate more on "how high the current is, that flows into the module." - what are you trying to do?
Are you using the input resistance to measure the current?
When it is ON, the input resistance if more than 100G ohm.
Santhosh
Soliton Technologies
(1,195 Views)
Thank you for the answer.
What I want to know is how much the voltage measurement "loads" my test setup. That is, how much power flows into or through the meter. I would have found the 100GOhm here as well. However, I was not sure, because this value seems to me to be very high.
(1,185 Views)
Yes, it should be high because of the OPAMP buffer at the input and thereby does not load the signal.
Santhosh
(1,173 Views)