06-25-2019 04:35 PM
Hi,
Are the commands to get the Relay Position (both Swtich and DAQmx deprecated) done by software (last command sent) or is there a physical verification?
Can that be used to verify if a relay is fused or defective or requires wetting?
Links to Get Relay Position:
http://zone.ni.com/reference/en-XX/help/370469AP-01/lvdaqmx/mxswitchgetrelayposition/
http://zone.ni.com/reference/en-XX/help/375472H-01/switchviref/niswitch_get_relay_position/
Thanks
06-26-2019 05:05 PM
Hello,
Which modules are you trying to test out? Are you having an issue with a specific module?
06-27-2019 08:26 AM
@Niki.B wrote:
Hello,
Which modules are you trying to test out? Are you having an issue with a specific module?
I'm not having issues, however I have a test requires the verification that the circuit is physically complete:
Context in particular: Short the low side of a differential signal (floating ~0V) to ground through the FIU card. We are trying to avoid extra circuitry to detect that the short is physically active.
Notes:
I expanded the question for other use of the commands.
My modules are the PXI-2510 and PXI-2569.
06-28-2019 10:02 AM
Hello,
The commands for the Get Relay Position VIs only provide a software verification, there is no hardware/physical verification of the connection. The physical verification will require extra circuitry.