07-04-2013 08:27 AM
I was calibrating a PXI 5124 with a NI-PXI-1042 when a error ocurred:
Error -1074135024 occurred at niFgen Initiate Generation.vi at step Verify Main Analog Path OffsetInvalid value for parameter or property. Device: PXI1Slot4 Channel Name: 0 Recommended Property to Modify: Offset Value: 1.750000 Maximum Value: 1.500000 Minimum Value: -1.500000 Conflicting Properties: Gain Value: 3.0 Status Code: -200269 Possible reason(s): IVI: (Hex 0xBFFA0010) Invalid value for parameter or property.
Calibration | As Found | As Left | |||||||
Gain | Test Value | Low Limit | Reading | High Limit | PassFail | Low Limit | Reading | High Limit | PassFail |
12 | 60000, V | 59485, V | 59996, V | 60515, V | Passed | 59785, V | 59996, V | 60215, V | Passed |
12 | -60000, V | -60515, V | -59990, V | -59485, V | Passed | -60215, V | -59990, V | -59785, V | Passed |
10 | 50000, V | 49570, V | 50001, V | 50430, V | Passed | 49820, V | 50001, V | 50180, V | Passed |
10 | -50000, V | -50430, V | -49991, V | -49570, V | Passed | -50180, V | -49991, V | -49820, V | Passed |
7 | 35000, V | 34697, V | 35004, V | 35303, V | Passed | 34872, V | 35004, V | 35128, V | Passed |
7 | -35000, V | -35303, V | -34989, V | -34697, V | Passed | -35128, V | -34989, V | -34872, V | Passed |
5 | 250000, V | 247825, V | 250045, V | 252175, V | Passed | 249075, V | 250045, V | 250925, V | Passed |
5 | -250000, V | -252175, V | -249890, V | -247825, V | Passed | -250925, V | -249890, V | -249075, V | Passed |
3.5 | 175000, V | 173463, V | 000000, V | 176537, V | N/A | 174337, V | 000000, V | 175663, V |
N/A |
What is the problem and how to fix it?
Thanks for the help.
07-08-2013 06:43 AM
Hi Jhsilva,
Apparently, your offset value has to be beetwen -1.5 and +1.5.
So, you have to decrease your offset value.
Best regards
Aurélien Corbin
National Instruments France
07-08-2013 08:30 AM
Yes, i did understand that, but the question is how?
I cannot change the value of offset without actually change the Calibration Executive procedure correct?
Or is it possible to do that in some other way?