12-15-2017 07:22 AM
I am using a Fluke 5500 Calibrator and the CAL-4353 to attempt the calibration of some NI PXIe-4353 cards. I am trying to follow the calibration procedure, but am running into some road blocks. The first problem is I can't "calibrate" the unit. I go into NI Max and find my unit. When I right click on it to configure, it reads the CAL-4353, but everything is grayed out and I can't execute the calibrate function. The other problem that I am having is I don't know where the SC Express Calibration Accessory Channels VI are recalled from. The procedure states " Call the DAQmx Connect SC Express Calibration Accessory Channels VI. Configure the CAL-4353 by setting the connection to _tc_calibration and the physical channels to Dev1/ai0:31". Throughout the procedure it is wanting me to configure the CAL-4353 and I don't know how to do that. Any assistance would be greatly appreciated.
12-19-2017
04:54 PM
- last edited on
07-10-2026
02:47 PM
by
Content Cleaner
In order to see the calibration tab you must first create a task(https://knowledge.ni.com/KnowledgeArticleDetails?id=kA03q000000x11DCAQ&l=en-US)
The SC Express Calibration Accessory Channels VI its a VI in LabVIEW see https://www.ni.com/docs/en-US/bundle/ni-daqmx-labview-api-ref/page/menus/categories/measurement/daqm...
03-27-2018 02:46 AM
Update:
I have written the code, but it is not working as I hoped. I am getting an error at the out of the Calibration Accessory Channels VI (error -200112) and at the end of the code (error -50103). I posted it as a different question (Cal-4353 Calibration) because I think once that works out it will answer all my questions about the PXIe-4353. Once I get everything working I will share my findings.