Hi terryatsbe,
Before assuming that the DMM is at fault I would check to make sure that the signal is still good on the other side of your 2527 multiplexer. In order to do this I would simply probe the wires that connect your multiplexer to your 4071 with an external DMM that you know works properly. If the voltage level with this is correct and does not exhibit the drop you are seeing this tells you it is not the 4071 that is the issue bit rather the 2527 multiplexer or some other external circuitry. If the signal getting fed into your 4071 from your 2527 multiplexer is correct then that means your DMM is not working properly.
There is a chance the issue is realted to your 2527 multiplexer module as well given the age of your hardware. In order to figure this out you will want to isolate all components of your system and see where the issue arises.
Best,
hfar2
03-17-2016 07:56 AM
Thanks for the ideas! Turns out that my "poking around" appears to have "fixed" the problem, as in "it's working better than in months". Coming from the point of view that
before getting my MSEE I did over 15 years in the home entertainment service business, I know for a fact that the problem will return. I just hope it is one of those
"really long term intermittent" situations..... Since the tester is "working", any further diagnose effort is futile.
One thing I tried was to input directly to the tester DVM the correct voltage that represented that which the tester would accept as "good". Since only one part out of 6
was being tested, and the multiplexer looks at all 6 test stations, the software error handling picked up inappropriate voltages at the unused test stations and initiated
an emergency shutdown, including the control computer. Although complicating [!] a diagnose process, I would argue that it was an entirely appropriate response
to provide another layer of operator safety.
When trying your suggestion to monitor the input with a DVM, I learned that all hand held DVM test leads are not created equal. The test leads I cannabalized
to implement your suggestion had pins just a crumb too small to connect with the input pins of the Fluke hand held I was using. I may be nearing 70, but I
can still be trained..... I hope retirement remains many years away!
Cheers!
Terry