PXI-2510 has no reaction while using NI VeriStand

Solved!
Go to solution

 

Hello all,

I want to use a PXI-2510 in NI VeriStand and I read a document named NI VeriStand Add-on - NI Fault Insertion Units at: http://zone.ni.com/devzone/cda/epd/p/id/6248

After deploy, I change state of controls in workspace, but PXI-2510 has no reaction.
 
The following is my step:

1. download the file-2510_add_on_nivs2011.zip

2. copy files in the folder named Custom Device llbs into <Public Documents>\National Instruments\VeriStand 20xx\Custom Devices

   copy files in the folder named Display Templates into <Public Documents>\National Instruments\VeriStand 20xx\Display Templates

3. open NI VeriStand and use PXI-2510 as a custom device.

4. deploy and open workspace. the default station of fault channels is pass through with dut channels. but when I change its station into an open state, PXI-2510 has no reaction.

Please help me with the problem.

Best wishes.
 

0 Kudos
Message 1 of 4
(5,905 Views)

by the way, name of custom device I use is as same as card name in MAX.

0 Kudos
Message 2 of 4
(5,900 Views)
Solution
Accepted by topic author MoonChen

If you use the console viewer (tools -> view console), do you see any error messages published to the screen?

Stephen B
Message 3 of 4
(5,887 Views)

Hello Stephen,

 

Thanks for your reply. I did what you suggest and find an error -200055. Then I search it in ni.com and find the problem-I use a dashes in name of my card in MAX.

 

well, I got this-http://digital.ni.com/public.nsf/allkb/E180150CE00F8A08862578ED00368929

 

also, thank you very much for help.

 

Moon Chen