Hello SiPHER,
The formulas for calculating absolute accuracy and the necessary constants are listed in the Specifications sheet for each board. For example, page 4 of the
6132/6133 Specifications document and pages 4 and 5 of the
628x Specifications contain tables with this information. The tables list a value for "Absolute Accuracy at Full Scale" which corresponds to the accuracy achieved when taking a measurement of the maximum voltage for a given range.
If you want to calculate the accuracy for a particular reading other than the full-scale voltage, you can use the formulas listed in the table for this purpose. I did a quick calculation for a voltage reading of 100 uV on each board, and I got the following results:
6132 - Absolute accuracy at 100 uV (1.25 V range): 358 uV
6289 - Absolute accuracy at 100 uV (0.1 V range: 13.6 uV
You can also use the
Accuracy Calculator on our website to find the same results for certain boards. Based on these results, I would recommend the 6289 for your application. I hipe this information helps.