04-29-2017 10:17 AM
Hi all, many posts talk about the possibility of signal generation by using a multifunctional DAQ, but it is not yet very clear to me. Just to fix ideas, refer to a generic mDAQ (6363 for example, I have not it, but I will :-)) if I want to generate a simple sine wave, what is the max frequency it can generate? 10k? 100K? 1MHz? Thank you !
04-30-2017 12:47 PM
To start with, the spec sheet for the device shows a max sample rate of 2.86 MS/sec for single-channel analog output. It lowers as you add more output channels, though not proportionally.
The next consideration is the # of samples per sine wave cycle. Too few makes a waveform that isn't very sine-like. I would generally aim for at least 20, though that still isn't especially smooth.
Finally, you need to figure out whether you will regenerate the same wave over and over or whether you will change it on the fly. If the latter, you need to account for the need to calculate and feed the task buffer with fresh data at your sample rate (on average).
With a 6363 type device, I'd venture you could regenerate a decent sine wave of up to about 100-150 kHz. That kind of ballpark.