question about DAQ sampling rate

Solved!
Go to solution
Highlighted

Hi,

 

As you can see the screenshot below, why aggregate multichannel sampling rate is lower than single channel sampling rate? 

ni.PNG

0 Kudos
Message 1 of 4
(207 Views)
Highlighted

@zyb1003 wrote:

Hi,

 

As you can see the screenshot below, why aggregate multichannel sampling rate is lower than single channel sampling rate? 

ni.PNG


You have a device that is muxed. The A/D conversion is shared between all of the channels. You would need a card that has a separate A/D for each channel, like the 6366.

 

mcduff

0 Kudos
Message 2 of 4
(160 Views)
Highlighted

Hi Mcduff,

 

You didn't get what I mean. Even though sampling rate is shared by channels, but in principle the aggregate rate should be same as maximum single channel rate. 

 

I searched similar question, seems reason is channel needs switching time for multichannels which sacrifice half sampling rate. 

0 Kudos
Message 3 of 4
(151 Views)
Highlighted
Solution
Accepted by topic author zyb1003
0 Kudos
Message 4 of 4