06-14-2006 11:45 AM
06-14-2006 12:48 PM
100V @ 100msec = POOF!
Manual states Max working voltage of +/-11V from AIGND, +/-25V overvoltage protection with DAQ powered on +/-15V powered off
What resolution do you require? This is a 16 bit card (65536 steps of discrete resolution). Worst case resolution would be using +/-10V input range. 20V/65536steps = 0.000305V/step.
0.000305v/20ohms = 0.000015A resolution (15uA). Not too bad, probably well below your noise floor. Things only improve if you use a lower voltage input range (+/-5, 2, and 1V available).
So with 0.5 short across 20 ohms sense resistor yields 10V drop, within spec of the DAQ card inputs. Can your setup allow a 10V drop across the resistor without affecting other things?
From the info you posted, I think you are OK. What do you think?
06-15-2006 02:15 AM
06-15-2006 10:34 AM
For 2uA accuracy, then off the cuff you need 0.2uA (200nanoAmp) resolution!
Personally I would suggest not using a DAQ card. Go with a lab quality DMM (Fluke, Agilent, etc) with proper specs on the current functions outfitted with a serial or GPIB interface. You wont have to worry about overvoltage issue with 0.5 amp short conditions.
06-15-2006 10:40 AM
06-15-2006 11:29 AM
Gogo-1962
Resolution is not equivalent to accuracy.
What type of voltage source is supplying the current you need to measure? Is it really capable of supply the burden voltage drop across the sense resistor? For example if you use a 200 ohm resistor with 0.5 amps of current, that is a 100V drop. If it is a 12V supply for example, you will never come close to generating 100V across the sense resistor.
You did not mention the need to measure other voltage channels simultaneosly, thus I threw out the DMM route.
Can you provide a clear description of the entire task and hand and the measurement requirements (sampling rates, AC or DC signals, voltage ranges, accuracies/resolutions required, etc)
06-15-2006 02:00 PM
The task in short is to characterize organic LEDs. One channel is the driving voltage in the range 0 - 60 V which through 1:10 devider is entering one channel of the board. The second channel is from a photodetector with 0 - 100 mV range. The third channel is the current through the LED. Typically the LED starts around 15 V with 5 mA current. But sometimes the silver paste that is used for contacts can give a short circuit that reportedly can produce no more than 0.5 A current. The power supply is a pretty big one with manual controls for the voltage and some 10 A capability. The current can be limited manually but not to a very small value. I do not see the relation between voltage drop on the resistor and power supply provided that the power supply can output 0.5 A. On 200 Ohms it is 100 V. So the measured signals are DC, the aquisiotion rate is once/second.
Some friends suggested using in the current channel a diode after e.g. 10 kOhm resistor to a 10 V supply, produced from the +12 V of the computer with a tsener diode and e.g. 200 Ohm resistor. So if the voltage excedes 10 Volts the diode will open and the current will flow through it and not through the board. But this probably will introduce noise, there is some leak current, and the system will need callibration. And I am not a hardware man. For any ideas I will be very thankfull.
06-19-2006 11:37 AM