Multifunction DAQ
output a floating signal from BNC-2110 and PCI6713
We are trying to control a current supply with analog output from BNC-2110 board and PCI6713 card. Our current supply requires a floating input, however we found the analog output only has a grounded reference, even it's already switched to 'FS'. Is there a way to generate a floating output? We appreciate any suggestions!
(1,157 Views)
You cannot change the AO into isolated. What current supply model?
There are isolated analog output cards like https://www.ni.com/en-us/shop/model/pxie-4322.html
Santhosh
Soliton Technologies
(1,140 Views)
Use a OP amp (or better two) to make the output float?
One amp and 4 identical resistors form a differential amplifier, the second buffer the output 'GND' or reference  to the diff amp reference.
But still migth depend on the actual voltage input of your current source. (Type?)
Greetings from Germany
Henrik
LV since v3.1
(1,104 Views)
If you don't want to roll your own, use a COTS approach such as this example (there are many other brands):
https://www.dataforth.com/dsca49
-AK2DM
(1,091 Views)
When using an external isolator, do consider its bandwidth and your signal frequency.
Santhosh
(1,086 Views)
Santhosh raises a valid point that I neglected to mention.
-AK2DM
(1,070 Views)