You don't say whether you are using NI-DAQ Traditional, or NI-DAQmx. You also don't say whether you are using Labview or C++ or CVI or..., so my suggestion here is pretty general, and I have aimed it at NI-DAQmx.
Use one of the onboard counter/timers to generate the digital pulse. You can trigger the pulse using
1) the same signal that triggers the analog input scan (if it's a digital signal)
2) If you are using an analog trigger for the input scanning, use /Dev1/AnalogComparisonEvent.
3) Use /Dev1/ai/StartTrigger.
The /Dev1/ai/StartTrigger will work even if the analog input scan is started programmatically.
You should be able to find an example of a triggered one-shot pulse generator in the examples included with NI-DAQmx, for whatever development environment you are using.