Interesting.
Mark- what is the input capacitance of the 6008/6009? The LM35's may get quirky above 50pF. Has anyone tried the decoupling/damping on pg 6 of the pdf supplied in the above post?
07-12-2006 12:44 PM
07-12-2006 12:45 PM
07-12-2006 12:56 PM
03-06-2007 06:18 PM
03-07-2007 09:35 AM
Hi Troy:
Have a look at Fig 8 in the app note:
http://cache.national.com/ds/LM/LM20.pdf
All I can suggest is to try to add a series resistor and parallel cap (try 470 ohm and 0.1uf). It looks like you have a match between the DAQ reading and scope. What happens when the DAQ is disconnected, does your sensor output rise above ~1V?. What voltage are you supplying to the LM20 and is supply ground connected to DAQ input ground? What mode is you DAQ set to (DIFF, RSE, NRSE)?
Sorry for all the queries
03-09-2007 11:22 AM
All I can suggest is to try to add a series resistor and parallel cap (try 470 ohm and 0.1uf). It looks like you have a match between the DAQ reading and scope. What happens when the DAQ is disconnected, does your sensor output rise above ~1V?. What voltage are you supplying to the LM20 and is supply ground connected to DAQ input ground? What mode is you DAQ set to (DIFF, RSE, NRSE)?