09-23-2005 09:49 AM
Is your analog input configured of single-ended mode, which is probably preferred over differential in this case.
Also, for kicks, have you tried connecting a ground to pin 1 of the 6008, though I do not think it is required, but still curious as to outcome.
09-23-2005 11:53 AM
Maxpower: Tested the program and got the same result as with the DAQ assistant.
Analogkiddigitalman: I have tested both RSE and differential, and it doesn't make any difference. I have connected pin 1 to ground also, and that didn't make any differense either. It's a stubbern 6008 I have....
09-23-2005 12:04 PM
09-23-2005 12:26 PM
09-23-2005 02:51 PM
09-23-2005 03:34 PM
09-23-2005 07:06 PM
09-24-2005 08:29 AM
05-09-2006 03:40 AM
Hi lads,
Im having the exact same problem trying to get my LM35 and usb 6008 DAQ to work...how did you eventually remedy this situation?
06-08-2006 10:38 AM
I've worked with the LM35 in the past (not with NI hardware), and have been looking today at the LM35/USB 6008 issue.
I've found two distinct ways to get it to work (return 'sensible' values, with no saturation problems).
1) Either power the LM35 from an external supply (I used 5V bench top PSU) and connect the output across the two pins of a differential input to the 6008 (so GND from the LM35 to AI- and Vout to the AI+)
2) Or power the LM35 from the 6008's 5V line (V+ to 5V line, GND to 6008 GND), connect the Vout to AI+ and a resistor between AI+ and GND, then connect AI- down to GND with another resistor of the same value. I'm currently using 120K, but values up to 220K should work.
I get the input saturating at 0.317V whenever I use the RSE mode on the unit. I'm assuming this is either due to the sensor not being balanced with the single ended Op Amp input, or some kind of floating voltage taking the signal outside of the common mode range of the instrumentation amplifier. These devices are often sensitive to the capacitance of the load they are driving, and this could be an issue with the inputs on the 6008, as they are fairly low impedance compared to, for example a DMM at 144K.
Anyway, if anyone's still working with this device and would like to try the above configurations, please post back with how you get on.
Best wishes,
Mark
Applications Engineer