10-26-2021 09:46 AM
I am trying to use USB 6009 to read thermocouple output in LabVIEW with AD595 as the amplifier. The problem is that I always get a voltage output of around 320 mV even if I heat the thermocouple. Maybe the output voltage changes slightly and I can't tell them visually, but it is absolutely not something like 10 mV/C.
The connection is based on the following figure. Pin 11 is connected to the 5 V supply of USB 6009. Pins 1, 4, 7, 13 are connected to the ground. Pins 8 and 9 are connected to AI0+ and another wire leads pins 1, 4, 7, 13 to AI0-. May I have some suggestions where I am wrong?
Many thanks,
Harry
10-26-2021 10:34 AM
You can ignore the AI0- and use the Analog Input in single-ended mode as your signal is also single-ended.
Now your connection will be just the Pin 8,9 to AI and you configure your DAQmx task to be RSE.
To ensure that your AD595 is working properly, power up the circuit and measure the voltage at Pin8, 9 using a handheld DMM.
BTW, I don't think you should connect Pin 1 to GND
10-28-2021 07:32 AM
Hi Santhosh,
Many thanks for your reply! I have tried it with your suggestion. It works perfectly! The only thing I didn't follow is that I still keep pin 1 connected to the ground.
When I disconnect pin 1 from the ground, the signals look like this:
But when I connect to the ground, the signals look better like this:
So I think this may be a better option. I also use a pull-down resistor to make it possible to measure lower temperatures.
Thank you very much for your help!
Best regards from UK,
Harry
10-29-2021 03:29 AM
Hi Harry
Are you sure that you want to use type K thermocouples in your application ? Unless cost is the overrideing consideration, you will get better stability from other types, e.g. N.
Obviously connecting pin 1 to common means that your thermocouple wires and junction have to be electrically isolated from the point where you want to measure temperature; this will introduce a responce lag.
Have you considered using a signal conditioning module? These can do all the hard work for you, and provide signal isolation. e.g. https://www.dataq.com/products/di-5b/di-5b47/di5b47n-15.html