Hi all,

 

I have the current set-up below:

- OMEGA pressure gauge connected to signal amp connected to USB-6002, finally connected to laptop (powered by AC adaptor) via USB

- Differential mode

- I am using a MatLab program to capture the data

 

I am seeing this step-wise function in the Volts (V) vs Time plot (s). Any idea why this is happening? The set-up is well grounded. I have attached a sample 10 sec data at 100 Hz.

 

Thank you in advance for your help!

 

Victor

test.JPG

 

 

 

 

0 Kudos
Message 1 of 4
(2,694 Views)

Addendum: This quantised effect is not present in RSE mode (Volts vs time (s))

 

test2.JPG

 

 

0 Kudos
Message 2 of 4
(2,683 Views)
Solution
Accepted by topic author victorkangnz

At a glance, it sure looks like A/D quantization.  Note that both the X and Y axes in your second posting have a significantly different scale than in your first.  The quantization is probably still there, it's just obscured by all the additional noise and the much wider view.

 

 

-Kevin P

ALERT! LabVIEW's subscription-only policy coming to an end (finally!). Permanent license pricing remains WIP. Tread carefully.
0 Kudos
Message 3 of 4
(2,667 Views)

Thank you very much, Kevin!

0 Kudos
Message 4 of 4
(2,628 Views)