Multifunction DAQ
Time out even though a measurement is made for pulse width
CT,
I need to fix a typo in my response.  It wasn't 1 sample per SECOND or 2 samples per SECOND.  It was 1 sample per CHANNEL and 2 samples per CHANNEL.  Setting the timing to Implicit/Finite with 2 samples per CHANNEL gave me a pulse width reading without the timeout error.
Message 21 of 24
(739 Views)
To VicL:
Welcome to the hardware forums.  Good job finding a workaround, let me help you out with a little more explanation about some things.
1. The best way to take 1 single measurement with DAQmx is to *never* call DAQmx Timing at all.  DAQmx Timing is used to reserve a buffer and set up a sample interval clock, neither of which is particularly useful if you're only taking one measurement.
2. As you found, DAQmx insists that if you *do* set up a buffered task, the buffer size needs to be at least 2.  This is a subtle hint at the same thing -- if you don't want to take at least 2 measurements then you also shouldn't be calling DAQmx Timing to configure a buffer and sample clock.
3. As you also found, when you *do* set up a finite task for 2 measurements, you're  still allowed to read only the 1st one.  Not typical, but it is allowed.
Go ahead and give it a try.  Take out the call to DAQmx Timing entirely and you should see that it works just fine that way too.
-Kevin P
Message 22 of 24
(728 Views)
Kevin P,
Thanks for your reply.  I had started my code without the DAQmx Timing VI and was getting the timeout error.  I tried implementing delays before and after starting the task.  I setup my hardware to send two pulses.  The two pulses would complete and then the timeout duration would finish running and I would get the error.  Nothing would get rid of the timeout error.  I even updated the DAQmx drivers with no change in results.  Then in my "Google" search I found this post, but still no resolution.  Only when I added the DAQmx Timing VI requesting 2 samples per channel did the timeout error disappear and the measurement worked as expected.
Thanks again,
VicL
Message 23 of 24
(714 Views)
To keep things clear for others who may come across this thread:
- what DAQ device were you using that gave you timeout errors whenever you didn't include the 2-sample DAQmx Timing call?    The shipping example "Counter - Read Pulse Width and Frequency (On Demand).vi" works fine for me with an X-series 6321 board.
- Can you post the code that produced timeout errors?  It'll be helpful to back-save to, say, LabVIEW 2016
- How did you manage the timing between when you requested the on-demand measurement vs. when the pulse(s) were actually generated?
-Kevin P
Message 24 of 24
(707 Views)
