Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

The sampling rate setting in DAQmxCfgSampClkTiming

Hi,
 
I would like to set up the sampling rate of PCI-6251with DAQmxCfgSampClkTiming(*taskHandle,"",SampRate,DAQmx_Val_Rising,DAQmx_Val_FiniteSamps,samplesPerChan) for multi AI channels. The 6251 manual says the highest sampling rate for signle channel is 1.25M and 1M for multi-channel. Does that mean I have to set the SampRate to be 1M or it has to be divided by the number of channels (500K if I have two analog inputs?)?
 
Thanks
David.
0 Kudos
Message 1 of 5
(3,526 Views)
Hi David!  look  here 

It seems that you have to divide!

Graziano
0 Kudos
Message 2 of 5
(3,519 Views)
Note: I refer to Ayman  post...
0 Kudos
Message 3 of 5
(3,517 Views)
Hello David,

Graziano is right, it's 1MS/s divided by the number of channels in the scan list.

Regards,
Micaela
Micaela N
National Instruments
0 Kudos
Message 4 of 5
(3,508 Views)
Thanks guys!
0 Kudos
Message 5 of 5
(3,500 Views)