Simulated compactDAQ not working properly

So I'm doing the online course on using NI-DAQmx; however since I don't have the hardware I use the simulation guide provided here http://www.ni.com/tutorial/3698/en/#toc2. But when I test it using a VI, it doesn't work as intended (as attached in picture); in the picture since the analog output frequency is 500 Hz I should see a peak at 500 Hz in the frequency domain plot, but this isn't the case in the picture. Anyone know what's wrong? Maybe I'm using the simulation incorrectly?

0 Kudos
Message 1 of 2
(1,603 Views)

Simulated devices are quite limited. It looks like your VI is expecting to output a signal on cDAQ3Mod4/ao2 which is then input to cDAQ3Mod8/ai2. However, input tasks on simulated devices are always read back as a full-scale sine wave with 3% of full scale noise. There's no way to link an output task into an input task, e.g., no "virtual wire" to connect different inputs and outputs on simulated modules.

 

I think you'd have to use your VI with real hardware to get the expected results.

——
Brandon Streiff
ni.com/compactdaq · ni.com/daq