Reduce noise of a 1 to 9 VDC measurement by USB-6001
Hi, we use the SCC-AI03 module in a SC-2345 box to measure the voltage signal (1 to 9 VDC) from a pressure transducer.  We plan to replace the old setup with a USB-6001.  We only need to record the pressure from two transducers.  We found the signal from USB-6001 has higher noise than the old SC-2345/SCC-AI03.
We use LabView to run the measurement.  The setting is as follow:
-- Terminal configuration: RSE
-- Timing settings: Model N samples, Samples to read 100, Rate (Hz) 1K
The SCC-AI03 has f0=10kHz.  How can we improve the performance of the USB-6001 measurement?  Thanks a lot.
(300 Views)
Please share data on the noise you captured.
Note that 6001 is an entry-level DAQ and not comparable to high-end X-series DAQ performance.
What DAQ were you previously using with the signal conditioning box?
Santhosh
Soliton Technologies
Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution
(286 Views)
Hi santo_13,
The DAQ currently used is 6036E.  I attached signal collected by USB-6001 and 6036E.  The timing is set as 100 samples and 1kHz.  We measured for 1 minute and get 600 data points.
We tried the USB-6001 is because we have one available.  Our budget is a little tight for this project.  Do you think a USB-6003 or USB-6421 can reduce the noise?
We use the single-ended connection on USB-6001.  Do you think the differential can reduce the noise?
Do you think a grounded enclosure help to reduce the noise?
(262 Views)
Accepted by Bobo31
it depends on what kind of noise you're dealing with
please read Field Wiring and Noise Considerations for Analog Signals - NI
Santhosh
(251 Views)
Hi Santo_13,
Thank you very much.  I am sharing the information with colleagues.  We are trying to reduce the noise of this DAQ setup following the instruction, and will use it as a guideline for the other projects.
(224 Views)
