12-30-2006 01:15 PM
01-02-2007 09:20 AM
Hi David,
What you are describing is expected behaviour for this device.
The 6120 has a maximum sample rate of 800kS/s per channel (1MS/s when operated in warp mode). This is defined in the product specifications and stated on the product page.
01-02-2007 12:08 PM
01-03-2007 03:10 AM
Apologies David,
I thought you were talking about the rate, clearly I need to ramp up after the Christmas lull!
01-03-2007 08:15 AM
Neil and Sarah, Thanks for the reply. I have posted most if not all of the pertinent exceprts from my code below. I have made comments in a different color to explain. It seems to me that there is something in the DAQmx drivers that limits the number of samples to 1.00M. Else it would seem the number of samples should be something close to a power of 2.
Neil, note that the issue of whether the number of samples is odd or even did not seem to affect anything. I used 1000001 samples in my post because the 'break point' was between 1000000 and 1000001. It finding that break point I used several other numbers of samples over 1000000, including (usually) even numbers. They all failed.
As you can see, I am using VB withing VS2005. As stated in my original post I am using DAQmx 8.3.0. I have recently gone through the agony of downloading 500M for DAQmx 8.3.1 but have not yet gone through the agony of a few hours of uninstallation and reinstallation time. My experience has been that every time I re-install DAQmx something breaks and it takes eons to fix but that was when I had to use traditional DAQ along side DAQmx because PXI 6120 modules were not supported in DAQmx and PXI 6123 modules were not supported in traditional DAQ.
A few notes:
My application requires producing a DO waveform that must be synchronized to AI sampling.
I use one 6120 module to produce the DO waveform and to be the master clock for AI sampling for all modules.
Producing the waveform and sampling the result requires setting up AI, DO, and Counter tasks.
The code below shows configuration and setup for all of AI, DO and Counters.
I have two different DAQ systems. One is based on PXI 6120 modules and the other is based n PXI 6123 modules.
Therfore, you will see some code options for each of the two systems.
As a side issue, at first I tried using DAQmx with multiple AI modules assigned to a single task.
See http://forums.ni.com/ni/board/message?board.id=232&message.id=3671
That proved to be much slower than using a single task per module and was never resolved so
I am using a single task per AI module.
'--------------------CONFIGURE DAQ--------------------
Call Configure()
See the entire subroutine below.
That subroutine does most of the set-up of the DAQ.
There are lines of code in that task that are commented out -- they are left there as notes.
'-----------------COMMIT DAQMX TASKS-----------------------------
'Commiting significantly speeds up AI task and can't hurt others
For k = 0 To mTEMData.DataPoint.AcqParams.nModules - 1
mAITask(k).Control(TaskAction.Commit)
Next
mDOTask.Control(TaskAction.Commit)
mCntrTask.Control(TaskAction.Commit)
'Start-------------------------
mDOTask.Start() 'Start DO, set for continuous output
DO is started first but is set to trigger from the first pulse from the counter
that also starts AI. So the Start command here only enables the actual start that
happens when the counter is started.
'---------------------Initial start-----------------
mAITask(k).Start()
Next
mCntrTask.Start()
mCPUTime.Time = Now
'--------------------------WAIT FOR ADCs-----------------------------
This is the section of code that hangs when the number of samples
exceeds 1M. See the commented lines in subroutine Configure
to see the conditions in which this section of code hangs and in which it
does not.
My.Application.DoEvents()
Do
If mAITask(0).IsDone() Then
Exit Do
Else
System.Threading.Thread.Sleep(1)
End If
Loop
xxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxx
Private Sub Configure()
Dim k&, s$, n&, iCh&
Dim x#
Dim mChan As AIChannel
'Tasks----------------------
mDOTask = New Task
mCntrTask = New Task
'AI chans-------------------------
iCh = 0
s = ""
mAITask(k) = New Task
'AI channels, input, type, gain
For n = 0 To mTEMData.DataPoint.AcqParams.nChannelsPerModule - 1
s = mDeviceName(k) + "/ai" + CStr(n)
x = 10.0# / mTEMData.DataPoint.AcqParams.Gains(Me.LogicalChannel(iCh))
Dim TermConfig As AITerminalConfiguration
#If DaqSystem = "NAVEOD" Or DaqSystem = "SERDP" Then
This statement is for 6120 modules
TermConfig = AITerminalConfiguration.Pseudodifferential
#ElseIf DaqSystem = "CRREL" Then
This statement is for 6123 modules
TermConfig = AITerminalConfiguration.Differential
#End If
mChan = mAITask(k).AIChannels.CreateVoltageChannel(s, "", TermConfig, -x, x, AIVoltageUnits.Volts)
These statements work only for 6120 modules
mChan.LowpassEnable = True
mChan.LowpassCutoffFrequency = 50000
'mChan.LowpassEnable = False
#End If
iCh = iCh + 1
Next n
Next k
'AI timing
If k = 0 Then
s = "OnboardClock"
This statement was tried in an attempt to get around the difficulty
I am having with more than 1M samples. I was postulating that the
difficulty was with a counter overflow, counting at the rate of 50 MHz
or 20 ns.
's = "100KHzTimebase" 'Causes error in task
Else
s = "/dev1/ai/sampleclock"
End If
n = mTEMData.DataPoint.AcqParams.NPtsPerInBlock
These statements were used in debugging to narrow down the source of
the hangup. Note that the code runs fine for n=1000000 but that
it hangs (or stalls) for anything larger. It does not seem to
be dependent on whether n is even or odd.
'n = 1024 * 1024 / 2 + 1 'OK
'n = 1000000 'OK
'n = 1020000 'Stalls
'n = 1010000 'Stalls
'n = 1002000 'Stalls
'n = 1000000 + 16 'Stalls
'n = 1000000 + 8 'Stalls
'n = 1000000 + 1 'Stalls
For this statement, Delt is 1.25 us for 6120 modules.
mAITask(k).Timing.ConfigureSampleClock(s, 1.0# / mTEMData.DataPoint.AcqParams.Delt, SampleClockActiveEdge.Rising, SampleQuantityMode.FiniteSamples, n)
'AI reader
mAnalogInReader(k) = New AnalogUnscaledReader(mAITask(k).Stream)
Next k
As a side issue, I would like to know if there is a faster way to gain
access to analog samples that should already be in memory.
It seems as if the reader is copying unscaled analog samples from its own buffer
to a new buffer. It takes on the order of 10 ms to read samples from 7 PXI6120 modules using the
UnscaledAnalogInReader. It seems I should be able to gain almost instant access to a buffer
that is already in memory.
01-04-2007 11:34 AM
01-04-2007 02:37 PM
[NOTE TO WEB SITE MANAGER -- I just spent a good long while composing a message. After clicking Post, I got a screen that said Authentication Failed. My entire message was gone and seems to need to be re-generated. It is quite frustrating.
Neil,
Thanks for the reply, this one will be be shorter.
My application goes like this.
The DO Waveform is driven by the Counter and the AI Sampling is triggered, but not timed by the Counter. It goes like this,
The DO Task is started but nothing happens because there are no counter pulses.
The AI Tasks are started but nothing happens because it has not been triggered.
The counter task is started. The first counter pulse triggers AI and all of the counter pulses provide timing for the DO waveform.
The AI Tasks and the Counter/DO tasks should be independent once the AI tasks are triggered. The AI tasks and the Counter/DO tasks should proceed to completion independently, even if their total times are different. The AI uses internal timing for AI sampling.
Typically a data block is 0.1 seconds. In each data there are 80,000 AI samples per channel (i.e. NPtsPerInBlock = 80000). Typically there are 36 counter pulses per in block, thus there are 36 DO states (actually 4 states repeated 9 times, such as in nine cycles of a repeated waveform). Thus NUpdates = 36 and the counter frequency is 36 pulses per 0.1 sec or 360 Hz.
In my application, a data block can be 0.3333, 0.1, 0.3, 0.9, or 2.7 seconds. The application works fine for all block times except 2.7 sec. The failure at 2.7 sec is what led me to discover that the AI tasks failed when asked to acquire more than 1,000,000 samples. Note that for 0.9 sec, there are 720,000 AI samples per block and at 2.7 sec, there are 2,160,000 samples per block.
r/e your question about the code
n = NUpdates()
mCntrTask.Timing.ConfigureImplicit(SampleQuantityMode.FiniteSamples, NUpdates)
The statement n = NUpdates was added during debugging to make it easier to check the value of NUpdates. As written it does not change the execution of performance of the code.
r/e DAQmx
I would like to see an optional DAQmx call that allows the user to provide the memory buffer to DAQmx. There are already calls that allow the user to override the size of the buffer (DAQBuffer.InputBufferSize()). It would seem easy to add a Method such as DAQBuffer.SupplyInputBuffer( byRef theUsersBuffer() as short). Perhaps your DAQmx software people would consider adding such a call. It would not complicate any user's life unless they chose to use that call. When choosing that call, the user becomes responsible as you say to determine whether data in the buffer are valid.
Dave George
01-05-2007 04:54 PM
01-08-2007 01:03 AM
Neil,
I have made several more tests in accordance with your last post. Here I will speak of two systems. System 1 is a system with Seven 6120 modules. System 2 is a system with Two 6123 modules.
I installed DAQmx 8.3.1 in System 1. It made no difference.
I tested any difference between Task.IsDone() and Task.WaitUntilDone(TimeOut). There was not difference.
As I am writing this message, it has occured to me that I could try using AnalogUnscaledReader(-1) and setting ReadAllAvailableSamples to false. I just made a brief test and it looked to function OK. It's too late tonight to try a full fledged test but tomorrow I will.
Here is some more information for you and a short program you can use that should reproduce the error. It has now turned up for me in two different systems. That fact that it is exactly 1,000,000 samples just sounds to me like a software test that is failing -- one of your software gurus may know exactly where it is.
System 1:
a. In MAX and In VB.net: Cannot set up multiple modules in one task. Causes Error 200072 Duplicate channels inthe list of physical channels are not supported by this device.
b. In MAX: One module: Can collect 2,000,000 samples per channel (4 channels)
c. In VB.Net, fails when 1,000,001 samples or more per channel, for either one module or two.
System 2:
a. In MAX: Can set up multiple modules in one task.
b. In MAX: Can collect 2,000,000 samples per channel (16 channels)
c. In VB.Net, fails when 1,000,001 samples or more per channel. Did not test just one module.
Below is a short program that fails in System 1. It hasn't been tested in system 2.
Public Class Form1
Private Sub Button1_Click(ByVal sender As System.Object, ByVal e As System.EventArgs) Handles Button1.Click
Dim mAITask(6) As Task
Dim mAnalogInReader(6) As AnalogUnscaledReader
Dim k, n As Integer, s As String, x As Double
Dim nModules As Integer = 2
'Reset all DAQ modules
Dim mDevice(nModules - 1) As DAQmx.Device
Dim mdaqsystem As DaqSystem
mDaqSystem = DaqSystem.Local
For k = 0 To nModules - 1
mDevice(k) = mdaqsystem.LoadDevice("dev" & CStr(k + 1))
mDevice(k).Reset()
Next k
'Set up tasks, one per module
For n = 0 To 3
s = "/dev" & CStr(k + 1) & "/ai" & CStr(n)
x = 10.0#
Next n
Next k
'AI Timing
If k = 0 Then
s = "OnboardClock"
Else
End If
'n = 1000 'Test
n = 1000000 'OK
n = 2000000 'Stalls
mAITask(k).Timing.ConfigureSampleClock(s, 800000.0, SampleClockActiveEdge.Rising, SampleQuantityMode.FiniteSamples, n)
'AI reader
Next k
'AI Commit and Start Start Task(0) last because it controls timing for other tasks
For k = nModules - 1 To 0 Step -1
mAITask(k).Control(TaskAction.Verify)
mAITask(k).Start()
Next
'Wait for AI done
Try
mAITask(0).WaitUntilDone(10000)
MsgBox("ADC's read")
Catch ex As Exception
MsgBox("AI read failed")
End Try
'Read AI data
Dim mData(,) As Short, mData2(,) As Short
mData = mAnalogInReader(0).ReadInt16(n)
mData2 = mAnalogInReader(1).ReadInt16(n)
End Sub
End Class
01-08-2007 07:41 PM
Neil,
I forgot to mention in my post that:
When Task.IsTaskDone() or Task.WaitUntilDone(timeout) is called and doesn't return (i.e. the root of the problem)
YES: the data is present in the buffer and when AIReader is called, the data appears to be correct. I haven't been able to check that waveforms are synchronized but I suspect they are.
Therefore, it is beginning to appear that the problem is within IsTaskDone or WaitUntilDone. It just seems that some programmer, some time, just believed that no one would ever want to take more than 1,000,000 samples in one acquisition block, so he wouldn't check beyond that. But so much for a wild-a__ guess.
Dave