03-09-2013 01:31 PM
Hello,
my PXI-6289 self calibration fails. I checked it with DAQ Diagnostic Utility 2.0:
-------------------- Testing Analog Input... -------------------- Testing AI Ground Reference... (PASS) Testing AI Voltage Reference... (FAIL) ************ Test FAILED for the following reason ************ Analog Input was not within tolerance for reference voltage. Expected to read: 6,950000 Actually read: 0,187908 ******************************************************* -------------------- Testing Analog Output... -------------------- Testing AO Finite Sample Clock... (PASS) Testing AO Finite Sample Clock Frequency... (PASS) Testing AO Single Point... (FAIL) ************ Test FAILED for the following reason ************ This test failed to output 1,000000 Volts on PXI1Slot5/ao0. The actual value written was 0,034417. *******************************************************
I had an issue with 12V supply rail - fuse F5 blew. I changed it, but the same problem still exists. I measured voltages on L1 through L8, are they OK?
L1: 0.7 V
L2: 20 V
L3: -20 V
L4: 38 V
L5: 12 V
L6: -12 V
L7: 20 V
L8: -20 V
What else should I check?
Thank you
Martin
03-18-2013 07:59 AM
Hello Martin,
Just to be sure - have you disconnected all wires from the card when performing self-calbiration and running the Diagnostic utility?
What are those voltages you've measured?
Thanks,
03-18-2013 01:00 PM
Hi Jakub,
yes, I disconnected all wires. The card is unfortunately malfunctional...
The voltages are measured on small inductors marked L1 to L8. Are the voltages correct? I would like to know if the problem comes from power supply of the board or whether it is bad analog part.
Thanks
Martin
03-20-2013 02:28 AM
Hi Martin,
Unfortunately I cannot check these voltages and tell you where the problem is. I suggest you to write an e-mail to techsupport@ni.com with link to this thread and a serial number of the card to start the repair process. Your e-mail will be redirected to proper local support, so you can use your local language.