Multifunction DAQ
Over heating of PCI 4472B
Highlighted
Dear Sir/Madam
I have 2 PCI 4472B cards on an ASUS mother board in my research lab. Due to lack of suitable cable it has not been used operationally up to now but the computer has been turned on for several times. Recently, I find that after few minutes of turning the computer on the temperature of four ICs after the analog part of PCI 4472B card, which are indicated by a red rectangle in attached figure, reaches to 63 Celsius with out using it for data acquisition. Self check, Self calibration and Test plan are done successfully by NI MAX.
Is this temperature is natural even in nonoperational condition?
Best Regards.
Hamed Kashani.
Message 1 of 8
(617 Views)
Highlighted
According to page 9 of the attached specifications, the operating temperature range is 0 to 50 °C and the storage temperature range is –20 to 70 °C.
Purchasing Agent for Apex Waves
Apex Waves
Message 2 of 8
(543 Views)
Highlighted
Hi Nichole,
Thank you so much for your reply. I had seen this document. This temperature range is related to environment but, I mean the temperature of the card itself.
Regards.
H. Kashani.
Message 3 of 8
(524 Views)
Highlighted
Unfortunately, that is not the normal working temperature of the module. Is the case being ventilated correctly? And if possible, can you find out the working temperature of your motherboard at that moment?
Thanks.
Message 4 of 8
(508 Views)
Highlighted
Hello, wanted to know if there are any updates on your side.
Message 5 of 8
(493 Views)
Highlighted
Yes, I measured the temperature of motherboard. It was about 25-30°C. Also, I tested the boards on other motherboard (Intel). Their temperature rises again. Chassis is ventilated correctly.
Sincerely.
Message 6 of 8
(482 Views)
Highlighted
Hello again,
If the temperature stays this way and the device is fully operational, then the problem must be in your case ventilation. I would recommend installing fans close to the card. The part that you squared on the photo are the ADCs, and whether you're acquiring data or not they heat up.
Thank you.
Message 7 of 8
(473 Views)
Highlighted
Hello
Thank you for your comment. I found a datasheet of ADCs IC "AK5393VS" as attached. At the first table of page 5. it is mentioned that ambient temperature (power applied) for this IC can be in the range of -10 to 70 °C. I try to have contact with AKM corporation about the problem. I will forward their reply to you.
Regards.
Message 8 of 8
(452 Views)
