Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

Our NI USB-6343 device is not outputting voltage values below -4 or above +4 volts.

Solved!
Go to solution

We are having trouble with our NI USB-6343 device.  Using AnalogOut, the output values supported are only -4 to +4 volts.  We cannot get any voltage values above or below to work properly.  Can someone please tell me what could be wrong, as we should be seeing values from -10 to +10 volts.

0 Kudos
Message 1 of 5
(272 Views)

Hi Brianbat, 

 

That behavior isn't right. How are you measuring the output? Do you have a load on the terminal? The card can only drive 5 mA so your voltage would probably start dropping if you are pulling more current. 

 

Best,

 

David F.

Technical Support Engineer

National Instruments

www.ni.com/support

0 Kudos
Message 2 of 5
(237 Views)

800 Ohm (or 1k) load?

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 3 of 5
(228 Views)
Highlighted

  

0 Kudos
Message 4 of 5
(224 Views)
Solution
Accepted by topic author brianbat

Thank you for the answers. The 50ohm impedence was causing some clipping on the top and low peaks.  We have since solved the problem by connecting a voltage follower circuit to overcome the impedence issue.

0 Kudos
Message 5 of 5
(223 Views)