Our NI USB-6343 device is not outputting voltage values below -4 or above +4 volts.
We are having trouble with our NI USB-6343 device.  Using AnalogOut, the output values supported are only -4 to +4 volts.  We cannot get any voltage values above or below to work properly.  Can someone please tell me what could be wrong, as we should be seeing values from -10 to +10 volts.
(2,049 Views)
Hi Brianbat,
That behavior isn't right. How are you measuring the output? Do you have a load on the terminal? The card can only drive 5 mA so your voltage would probably start dropping if you are pulling more current.
Best,
David F.
Technical Support Engineer
www.ni.com/support
(2,014 Views)
800 Ohm (or 1k) load?
Greetings from Germany
Henrik
LV since v3.1
(2,005 Views)
(2,001 Views)
Accepted by brianbat
Thank you for the answers. The 50ohm impedence was causing some clipping on the top and low peaks.  We have since solved the problem by connecting a voltage follower circuit to overcome the impedence issue.
(2,000 Views)
