Offset at PCI 6120 if 100kHz antialiasing input filter is active

The PCI 6120 shows a range dependent offset voltage if the internal 100kHz antialiasing filter is activated. For the +/-1V range the offset is about -11mV, for the +/-10V range the offset is about -138mV, both measured with shorted input. It is the same offset for AC and DC input coupling. Each input channel shows a different offset, but all are far above the specification.
If the filter is turned off, everything is fine. I have performed a self-calibration and I observe a slight change in the offsets after the calibration. It looks like there is no offset adjustment for activated filter in the self-calibration routine.
Any idea?
LabView 2010SP1, Windows7
0 Kudos
Message 1 of 9
(3,944 Views)

Hello hejoba,

which driver do you use with the pci 6120.
If you are using NI-DAQ 6.9.3f3 and earlier there is a software bug 

If you are using a newer driver please let me know which version and which development environment do you use LabVIEW, LabWindows/CVI the NI-DAQ API.

Have a nice day

Benjamin

0 Kudos
Message 2 of 9
(3,932 Views)
Dear Benjamin,

thank you very much for your reply. Here are my system specs:

LabView 8.2.1 prof. dev. syst.
NI-DAQmx 8.5.0f3

The offset does not appear if I use SignalExpress as a simple data viewer, but with SignalExpress there is no chance to activate the 100kHz filter. If I convert the SignalExpress measurement into LabView code and activate the input filters by use of the DAQmx channel property, the offsets are back.

I have also checked different sampling rates, but its the same. Here are the offset values of all 4 input channels at +/-10V range, AC input coupling, and activated input filters:

ch1: -128mV
ch2: -1mV
ch3: -18mV
ch4: -43mV

The PCI 6120 manual notes, that the DC offset might degrade by 8 LSB with activated filters. 8 LSB at +/-10V range are about 2.4mV. If I add this to the specified maximum offset without filters of 1.7mV, I end up at 4.1mV maximum offset with activated filters. So, only channel 2 meet this spec.

The last external calibration is from May 2007 (that is right before I purchased the card). The calibration temperature was -128°C. Is that correct?

Thanks in advance!
LabView 2010SP1, Windows7
0 Kudos
Message 3 of 9
(3,924 Views)
Hello hejoba,

i think you can eliminated the issue by doing a self-calibrate in Measurement and Automation Explorer -> Select the PCI 6120 and press selfcalibrate.

Please let me know if the problem consist

Best Regards

Benjamin
0 Kudos
Message 4 of 9
(3,876 Views)
Dear Benjamin,

I have tried the self-calibration several times, but it does not help. There is only a slight change of the offsets after self-calibration. If have tried the self-calibration with open and shorted inputs without success.

Regards, hejoba
LabView 2010SP1, Windows7
0 Kudos
Message 5 of 9
(3,872 Views)
Hello hejoba,

i will check this her at National Instruments with a PCI 6210.
Do you have a signal connected to the input ?
Or do you have a offset if you connect the input to ground ?

Best Regards

Benjamin
0 Kudos
Message 6 of 9
(3,836 Views)
Dear Benjamin,

the offset is present with open input and with shorted input (i.e. shorted to ground). If I connect an AC or DC signal to the inputs, the offset is there also. So, it really looks like the self-calibration ignores the offset caused by the input filters.

Regards, hejoba
LabView 2010SP1, Windows7
0 Kudos
Message 7 of 9
(3,835 Views)

Dear hejoba,

could you please download the new DAQmx 8.6 NI-DAQmx 8.6 - Windows 2000/Vista x64/Vista x86/XP because there is a known issue with the antialiasing input filter and DAQmx 8.5

Regards
Benjamin

0 Kudos
Message 8 of 9
(3,762 Views)
Dear Benjamin,

please excuse my late reply but I was really busy. With the new Software everything works fine. The offsets are within the specs (even much better) now. Thank you very much for your help.

Regards, hejoba
LabView 2010SP1, Windows7
0 Kudos
Message 9 of 9