Multifunction DAQ
Noise problem PCIe-6259
I'm having some issues with a PCIe-6259. The problem is with all the analog inputs, there is noise in all of them. In order to remove signal conditioning from the problem I've disconnected all the signals, and I was hope of a solid 0V, but I instead of this there is a 0,02 Vpp oscilating signal (see attached screen shot). Is it possible that the PC is generating the noise? Are there PC minimum requirements in order to avoid noise???
Mariano
(4,679 Views)
Disconnecting the inputs will not give you a 0V reading. What you are seeing is normal with floating inputs. If you want to read 0V, you need to GND the inputs.
(4,677 Views)
Ok, see the attached file. That is a AA battery conected to channel 0 in my SCB-68 block with a shielded cable to the PCIe-6259. With the input range the resolution is 0,000320 V, and the signal has a 0,02 V pp. That is noise I'm sure. Can be the PC?
Mariano
(4,659 Views)
You are doing a differential measurement. What pins on the SCC-68 do you have the battery connected to?
(4,643 Views)
Dennis,
I made a mistake, the battery wasn't conected properly for a differencial measurement.
I corrected that conecting the positive terminal of the battery to AI0, the negative terminal to AI8 and I also wired AI8 with AIGND, I think that's the correct differencial conection for a floating signal source as a battery. I ran the test panel and I saw a rms noise of about 0,002 V. This is 10 times better that in previous situation, but It's still to much taking in count that for a +/- 10 range the resolution is 0,000320 V. The battery is wired to the block terminal with a short wire (1 ft), and I look at the signal with high resolution DMM and the signal is perfectly stable in 1,517 V
Thanks for your interest.
Mariano
(4,594 Views)
Buenas Tardes
Siempre pudiste resolver lo del ruido.....yo tengo el mismo problema.....tengo un ruido que no puedo disminuir de +/- 5 mV en mis entradas analogicas de una tarjeta PCI6320? Que podria hacer para reducir este ruido a menos de +/- 1 mV?
(3,258 Views)
