01-13-2010 11:59 AM
I have the problem that my NI cDAQ-9172 is not more viewed/recognized in MAX. But this hardware is well recognized by Windows (2000 SP4) in the Device Manager under Data Acquisition Devices and also under the USB Root Hub (it is connected with a USB cable). "This device is working properly"
This device was working well before. I just connected last week a light source device (not from NI) with a USB connection + installed its driver. This lignt source is not more in the lab and so not more connected to the computer. Today it is the first time that I try to work again with my NI hardware.
I restore the MAX database but nothing changed.
What can I do to "force" MAX to find the cDAQ?
I have a FW camera connected and this one is still recognized and working with MAX. The problem seem to be connected to the USB connection.
Thank you in advance for your help.
Patrick Queeckers
Solved! Go to Solution.
01-13-2010 09:13 PM
01-14-2010 06:41 AM
Solution found.
With the help of an engineer of NI from Belgium, we found the solution: the NI Device Loader was not running !!!
To started it:
After started this service, MAX found the NI cDAQ hardware and it works properly.
But after a reboot of the computer, this service was again not started (it is in automatic mode) and the hardware was listed in the MAX but not working.
I changed the properties of the service "NI Device Loader" in the recovery tab, I put the "First failure" parameter to "Restart the service". After that is it working after a reboot.
01-14-2010 01:30 PM
Hi Patrick,
That sounds like a reasonable workaround, but NI Device Loader should not stop running like that. Could you open Event Viewer and check the System log for events regarding NI Device Loader or NI-PAL? Look for events that have "nipalk" or "nidevldu" in the Source column, or where "Service Control Manager" is in the Source column and the event description mentions "NI Device Loader".
Brad