Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

NI 9862 in cDAQ 9172 " unsupported module" error

Solved!
Go to solution

Hi

I just purchased a NI 9862 CAN module to use in my cDAQ 9172 but the module is not recognized in MAX. I use XNET 14.5 (because I have Labview 2011 this is the latest supported version). According to the documentation NI 9862 is supported by XNET14.5

 

What to do?

NI support report attached

 

Regards

Mads Nygaard

Dinex

 

MAX Configuration Report:

MAX Configuration Report

02-01-2019 10:18:08

MAX Summary

0 Kudos
Message 1 of 3
(3,004 Views)
Solution
Accepted by topic author MadsNygaard

The 986x modules are not supported in the cDAQ-9172 (Legacy). See the C Series Module Compatibility Chart.

 

The cDAQ-9172 was the first generation of CompactDAQ chassis and lacks certain hardware features needed for newer C Series modules. The cDAQ-9178 is its direct replacement. (Or, if you're willing to rearrange things, the 986x modules will also work in the 9171s you have.)

——
Brandon Streiff
ni.com/compactdaq · ni.com/daq
Message 2 of 3
(2,988 Views)

Hi Brandon

Thank you for the reply. I hadn't found that overview in my search. 

Guess I need to make an investment 🙂  

0 Kudos
Message 3 of 3
(2,976 Views)