NI 9211 sampling rate

Solved!
Go to solution

Hi,

I am using NI 9211 together with cdaq NI 9717. I wonder what is the meaning of 14S/s "aggregate". As stated in the datasheet that the sampling rate for NI 9211 is 14S/s which also means the conversion time is 70ms per channel, but when I tested in the NI MAX by input "14Hz" in the timing setting , it shown with an error message: Error -200332 occurred at DAQ Assistant.

Possible Reason(s):

Specified sample rate is higher than the fastest rate supported by the device.

Property: SampClk.Rate
Requested Value: 14.0
Corresponding Value: 14.000002

Maximum Value: 7.142857

which make me curious about the maximum value on the sampling rate, shouldn`t it able to run 14S/s, instead of 7.142857S/s

Could anyone please explain the function of "aggregate" or suggest the corrective action. Thank you very much.

0 Kudos
Message 1 of 2
(846 Views)
Solution
Accepted by topic author GracelynnMa

Hi, the answer to your questions is explained on this KB: What "Maximum AI Sample Rate Single Channel and Aggregate" Values Means?
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000kIz9SAE&l=en-US