06-13-2011 11:16 AM
Like the title says, my analog output channels have a slight offset. Is there any way to calibrate them to get them to return to 0.
Thanks.
06-14-2011 03:32 PM
Michael,
Are you always seeing the offset or just at power-on. Your device has a Power-On state of +/- 200mV per the 6713 Specifications. This is further explain in What Is the Analog Output Power-On State of the PCI-6713?
If it is always occuring afte the power-on, then you may need to perform a calibration. A contributor to offset errors is the difference in the current temperature of the card, and the temperature that is was calibrated at. You can perform a 'self-calibration' on your device in Measurement & Automation Explorer to compensate for this difference. To perform a self-calibration: Open MAX>Expand Devices and Interfaces>Right click on your PCI-6713>Choose 'Self-Calibrate'>Click 'Next' then 'Finish' on the dialog boxes.
If you are still experiencing an offset after the self-calibration, then it may be time to calibrate your card. If you have the proper equipment, you can follow the directions in AO Waveform Calibration Procedure for NI-DAQmx. If you do not have this equipment, then either NI or a 3rd party can perform a calibration for you. Click here for more info on NI's calibration services.
Regards,