Multifunction DAQ
Measuring voltage from thermocouples
Dear all,
I have a doubt about reading voltages from thermocouples.
I've read many discussion on this forum on this topic (for instance http://forums.ni.com/t5/Multifunction-DAQ/measuring-millivolts-cDAQ-NI-9205/m-p/950562/highlight/tru... http://forums.ni.com/t5/Multifunction-DAQ/Thermocouple-channels-stay-differential-once-you-assign-th... http://forums.ni.com/t5/Multifunction-DAQ/Thermocouple-Noise-and-Calibration-SCB-68/m-p/214645/highl... but there is something that is not clear to me.
I'm trying to read voltages (in differential mode) and trasform manually these voltages into temperature from a E-type thermocouple directly connected to my SCB-68 terminal block.
In order to avoid errors I'm following this procedure:
a) manually do a short circuit on PIN ai1-ai9;
b) read voltages @ 1 kHz in order to compute the mean offset voltage (Voff);
c) connect my thermocouple to PIN ai1-ai9 and record voltages (Vmeas) while checking the temperature of the internal temperature sensor for CJC (Visothermal);
d) compute the value of the "corrected" voltage from the following equation: (Vmeas-Voff) + V_isothermal;
e) associate this value to a temperature thanks to the V-T E-type TC curve.
The things seems to be quite good at room temperature, I compute 24.4 °C while the CJC perform 25 °C, and maybe I can improve my software with a filtration in order to avoid the 50 Hz noise.
However I want to know if there is a method to do the Voffset via software and not manually.
Unfortunately I got a NI-PCI-6281 which seems not to support the OTD (open thermocouple detection) which should solve my problem.
Do anybody have some suggestion?
I need to perform all this procedure since I have to do measurements below -200 °C so that I cannot use the NIST polynomial functions and AI temp TC channels (I mean the authomatic temperature measurements of LabVIEW).
Thank you very much for your help.
Marco C.
(2,788 Views)
your measure mode is optimal but you can use also MAX configuration to do compensation and give in output (setting as well a custom scale) directly temperature information.
Have you ever performed a thermocouple measure directly by MAX?
Giuseppe.
(2,734 Views)
