I'm trying to figure out the best way to measure the internal voltages across an electrolyzer stack containing 10 cells. Voltage range will be 0-20V, with each differential voltage measuring 0-2V. I know that each module on a NI-DAQ (6218, 6210, 6009, etc) can not see voltages greater than 10V. If I use the AISense or ground the second board to the high voltage from the first 5 cells (~10V), will I damage the board or will that enable me to reach the 20V upper limit of the system. Additionally, I need to be able to measure the voltages simultaneously.
If there is a better piece of hardware or board for this topic, please let me know. I'm posting on the multi-function DAQ board because that's the hardware that I'm familiar(-ish) with and it's what we have in house right now.
Solved! Go to Solution.
04-16-2012 08:00 AM
I should note that I do know that I can use voltage dividers and condition the signals that way, I was just wondering if there is a way to do this without the voltage divider. Thanks.
04-17-2012 05:37 PM
Hi sbrawn,
I recommend reading the High-Voltage Measurements and Isolation DevZone.
If you are looking for a DAQ device that has a voltage range greater than 10 V, I recommend searching the product catalog here.
04-18-2012 02:38 AM
How about 5 INA2128 ?
This is an instrumentation amplifier already including 8 laser trimmed resistors and a input protection circuit and work up to 36V
Depending on the input range of your DAQ you could add gain (however might not be needed)
Use the DIP version in a socket, so if you blow one up is a quick replacement 😉