Measuring Coil Resistance with NI 9219
Trying to add attachments again.
Message 11 of 17
(2,561 Views)
For some reason the Add Attachment function is not working, will try later.
Message 12 of 17
(2,559 Views)
The Forum software can be picky about data types for attachments. Try .png or .pdf files or zip the files and attach the compressed archive.
Or use the Insert/edit image button (tree icon) in the toolbar above the text box for posting. That will insert your image directly into the text.
Lynn
Message 13 of 17
(2,552 Views)
Schematic diagram
Message 14 of 17
(2,549 Views)
Wiring Plan
Message 15 of 17
(2,548 Views)
Nicely documented!
The only thing I can think of at the moment is to question the timing of the opening and closing of contacts when the Position Select Relay (K1) operates. During the transition is the common terminal completely open to both connections for a short interval or possibly connected to both simultaneously (unlikely)?
If both contacts are open briefly and the stored energy has not dissipated in the arc at the contact by the time the other contact closes, a spike could occur.
Can you check the contacts with an oscilloscope to see if some transients are getting through?
The Position Select and Coil Select switches are drawn as center open (3-position) switches. It does not matter for the Position Select since only one position controls K1. If the Coil Select switch is really a 3-position switch, try setting the selection to the center (No Coil Selected) position before switching K1. Since the coil grounds are switched by the Coil Select Relays, de-energizing BOTH relays before switching K1 should eliminate the possiblity of a transient when K1 switches. If S2 is not a 3-position switch, just disconnect its common from +15 V before switching K1 as a test.
Lynn
Message 16 of 17
(2,537 Views)
Lynn,
Thank you, I will try your suggestions tommorrow when I get to work.
Message 17 of 17
(2,528 Views)