The 9237 is dedicated to bridge measurement.
The input voltage range according to spec is +-25mV/V , so if set to 10V exitation you have a ~+-250mV range you can (ab)use for voltage measurements. Just keep in mind, that the bridge measurements are ratiometric to supply voltage, (the ADC reference voltage is tied to the exitation voltage) . If the exitation voltage drops due to load conditions, the ADC value will follow while the voltage ratio of a bridge is still the same.
I don't know what values the 9237 reports (mV/V or simply V corrected by the nominal value..) and since the exitation is shared with the other channels, how big the resulting common voltage range for that channel is.